Dear Dr. Hagemann
I am estimating the depletion voltage to verify if there are any errors in detconfig.yaml. Additionally, I performed some tests: executing Could you provide some suggestions or enhancements on estimating the depletion voltage for a double-sided orthogonal strip detector using SSD? Setting the potential of all top electrodes to 1200V in detconfig.yaml may be too high. I would like to determine the potential setting based on the estimated depletion voltage. Below is my detconfig.yaml, hoping it will help you better understand my issue. |
Beta Was this translation helpful? Give feedback.
Replies: 2 comments 1 reply
-
Hi,
Both methods return quite similar results, so you can assume that your detector is fully depleted already at 1050V. |
Beta Was this translation helpful? Give feedback.
-
Of course, both of these methods require to recalculate the electric potential and weighting potentials. |
Beta Was this translation helpful? Give feedback.
Hi,
as you pointed out, the current implementation of estimating the depletion voltage assumes that the non-zero potential is only applied to one of the contacts.
There are some tweaks you could use to modify the detector such that you can use the SSD function:
Combine all contacts with potential 1200V into ONE contact, and all contacts with potential 0V into ANOTHER contact, so you end up with a detector that "consists" of only two contacts. Note that the following code is hard-coded and only works if the biased contacts are at potential 1200V and if the first contact is a biased contact and the second one is not: